OpenBCM V1.07b12 (Linux)

Packet Radio Mailbox

IW8PGT

[Mendicino(CS)-Italy]

 Login: GUEST





  
G8MNY  > TECH     21.02.17 12:47l 66 Lines 2031 Bytes #999 (0) @ WW
BID : 35714_GB7CIP
Read: GUEST
Subj: Measuring 1 of 3 Rs
Path: IW8PGT<IZ3LSV<IR1UAW<IQ5KG<I0OJJ<GB7CIP
Sent: 170221/1010Z @:GB7CIP.#32.GBR.EURO #:35714 [Caterham Surrey GBR]
From: G8MNY@GB7CIP.#32.GBR.EURO
To  : TECH@WW

By G8MNY                                           (new Mar 06)
(8 Bit ASCII graphics use code page 437 or 850, Terminal Font)
I recently bought an old LAB DMM (Datron 1061a) meter. It has an unusual
feature, the ability to measure one Resistor R1 in a triangle of 3 (R2 & R3)
while all connected.

ÚÄÄÄÄÄÄÄÄ¿
³    HighÃÄÄÄÄÂÄÄÄ¿
³        ³    ³  R2
³     Rx>³   R1   ÃÄÄ¿
³ DMM    ³    ³  R3  ³
³     LowÃÄÄÄÄÁÄÄÄÙ  ³
³        ³           ³
³   GuardÃÄÄÄÄÄÄÄÄÄÄÄÙ
ÀÄÄÄÄÄÄÄÄÙ

As the DMM has uProcessor you might expect the Guard line to be connected to
the Low then High & then disconnected, to give 3 values of Rx, then with 3
equations & 3 unknowns all the Rs can be calculated....

1/ Guard                  R1 + R2
   connected         Rx = ÄÄÄÄÄÄÄ
   to Low                 R1 x R2

2/ Guard                  R1 + R3
   connected         Rx = ÄÄÄÄÄÄÄ
   to High                R1 x R3

3/ Guard                  R1 + R2 + R3
   not               Rx = ÄÄÄÄÄÄÄÄÄÄÄÄ
   connected              R1 x (R2+R3)

But that is not the purpose of the Guard line in a sensitive measuring circuit.

To measure the resistance the DMM actually uses an accurate constant current
source across Rx & measures the voltage to get the R1 value. But with R2 + R3
in circuit this will give a wrong reading as some of the accurate current will
flow through R2 + R3. But on this meter the Guard line has another current
generator with its control arranged to make the Guard the same potential as the
Low...

ÚÄÄÄÄÄÄÄÄÄÄÄÄ¿
³        HighÃÄÄÄÄÂÄÄÄ¿
³          / ³    ³  R2
³        DVM ³   R1   ÃÄÄ¿
³          \ ³    ³  R3  ³
³  Ref I> LowÃÄÄÄÄÁÄÄÄÙ  ³
³       /³_ /³   No /³\  ³
³  ÚÄÄ<' ³_  ³  Voltage  ³
³ \³/  `\³  \³      \³/  ³
³ Stray I> GuÃÄÄÄÄÄÄÄÄÄÄÄÙ
ÀÄÄÄÄÄÄÄÄÄÄÄÄÙ

So as long as the Guard stray current generator can put the error current
through R2 the Reading of Rx = R1 & be correct.


See my tech bul "Lab Measuring DMM".

Why Don't U send an interesting bul?

73 De John, G8MNY @ GB7CIP


Read previous mail | Read next mail


 11.05.2024 09:07:40lGo back Go up